500 Volt Insulation Test

a) Ensure equipment is still plugged in and make sure the ‘on/off’ switch is in the ON position
b) Attach the wander lead to a metal part of the appliance- this should always be done but often it is forgotten.
c) Follow the on-screen instructions on the Metrel AlphaPAT or MetroiPAT until the test is completed
d) At the end of the test ensure the appliance is turned off.
When testing insulation on a class II appliance, ATTACH the wanderlead to a exposed metal part of the appliance.
NOTE: The MetroIPAT will always do a 250/500 Volt Insulation Test first if both a leakage and a 500 Volt Insulation Test are selected.
For the equipment to pass this test the reading displayed must be 10 Megohm or greater.
Please see the values below:
AS/NZS 3760
| 250 V DC | 1MΩ |
| 500 V DC | 1MΩ |
AS/NZS 3551- Appendix B 9.3
| Non IT | 10MΩ | 500V |
| IT | 10MΩ | 250V |